Hi all, I did the Boil and ice calibration tests for the probes today and here are my results:
Boil
Probe 1 - 209
Probe 2 - 209
Ice
Probe 1 - 32
Probe 2 - 33
For you more experienced guys out there, what would you program in for the offset values. Looks like a 3 degree difference on the high side as water boils at 212.
Is it close enough to just let it slide?